Reference Materials Setting Bidirectional Digital Bits> |
Instruments with Bidirectional Digital bits require you to set the direction prior to reading or outputting digital signals.
Setting the Direction of the Digital Bits
The directions of the Digital Bits on DI-71x Instruments MUST be set using the DATAQ Instruments Hardware Manager. View the DATAQ Instruments Hardware Manager Help file for more information.
The directions of the Digital Bits on DI-148 and DI-158 Instruments should be set using DigitalOutput. DigitalOutput is a 16-bit integer property.
Note: You may set the direction of the digital bits for Models DI-148 and DI-158 Instruments using the DATAQ Instruments Hardware Manager. This will write the direction directly to the instruments. However, each time the instrument is powered, the digital bits automatically reset to inputs requiring you to reset the bits each time.
Bit 15 determines whether the lower 8 bits are used for digital output or if they are used to determine the direction of the digital ports.
Bit |
0 |
1 |
15 |
indicates the lower 8 bits are for digital output. (0 for low signal, 1 for high signal) |
indicates the lower 8 bits determine direction (1 for input, 0 for output) |
Bits 8 to 14 are reserved and should be 0.
Bits 0 to 7 correspond to the digital ports on the hardware and determine direction if Bit 15 is set to 1 or determine a high/low level signal if Bit 15 is set to 0.
when Bit 15 = 0 |
when Bit 15 = 1 |
Port |
Bit |
0 |
1 |
0 |
1 |
D7 |
7 |
low |
high |
output |
input |
D6 |
6 |
low |
high |
output |
input |
D5 |
5 |
low |
high |
output |
input |
D4 |
4 |
low |
high |
output |
input |
D3 |
3 |
low |
high |
output |
input |
D2 |
2 |
low |
high |
output |
input |
D1 |
1 |
low |
high |
output |
input |
D0 |
0 |
low |
high |
output |
input |
To set all Digital ports to outputs you would use the following: DataqSDK.DigitalOutput = &H8000
To set all Digital ports to inputs you would use the following: DataqSDK.DigitalOutput = &H8000 + 255
To set only the second digital port to an input you would use the following: DataqSDK.DigitalOutput = &H8000 + 2
To specify digital bit directions on multiple DI-1110, DI-1120, DI-2008, DI-2108, DI-2108-P, DI-4108, DI-4208, or DI-4730 devices, to each direction configuration value add 256 times the number of connected devices with lower serial numbers or listed ahead of it in the DeviceID property.
Note: Bits corresponding to ports that do not exist on your hardware are ignored. To read Inputs after starting and stopping a scan, use the DigitalInput property concentrating only on the bits designated as Inputs. To read inputs during a scan, enable a digital channel in the scan list and look at the most significant byte of the 16-bit integer data scanned for that channel.
Use an 8 bit integer and the DigitalOutput property to send output signals. Bits programmed as inputs will ignore the command.
DataqSDK.DigitalOutput = 255 will output 1 on all digital outputs - only ports programmed as outputs will see the signal.
DataqSDK.DigitalOutput = 0 will output 0 on all digital outputs - only ports designated as outputs will see the signal.
DataqSDK.DigitalOutput = 2 will output 1 on D1 - as long as D1 is designated as an output.
To select among multiple DI-1110, DI-1120, DI-2008, DI-2108, DI-2108-P, DI-4108, DI-4208, or DI-4730 devices, to each output value add 256 times the number of connected devices with lower serial numbers or listed ahead of it in the DeviceID property.